Aller au contenu principal
How to Test the Valve System on a Fisher Paykel Neopuff
Matériel médical
Ni iFixit ni ce tutoriel de réparation ne sont approuvés par ou affiliés au fabricant de ce dispositif médical.
    • Set the gas supply to 5 L/min.
  1. How to Test the Valve System on a Fisher Paykel Neopuff: étape 2, image 1 de 1
    • Completely close the Peak Inspiratory Pressure control and the Max Pressure control by turning both knobs completely clockwise.
  2. How to Test the Valve System on a Fisher Paykel Neopuff: étape 3, image 1 de 1
    • Close the bleed valve and check that the gauge reads at least 60 cmh2O.
    • Set the gas supply to 15 L/min.
  3. How to Test the Valve System on a Fisher Paykel Neopuff: étape 5, image 1 de 1
    • Close the bleed valve and check that the gauge reads no higher than 80 cmH2O.
    • Set the gas supply to 10 L/min.
    • Close the bleed valve and turn the Max Pressure Relief valve until the manometer reads 40 cmh2O.
    • Check that the manometer needle rises and falls smoothly.
    • Gently rotate the Max Pressure Relief cover over the Max Pressure Relief knob.
  4. How to Test the Valve System on a Fisher Paykel Neopuff: étape 10, image 1 de 1
    • Reset the peak inspiratory pressure to 20 cmh2O and turn off the gas flow. Testing is now complete. If the F&P Neopuff/Perivent Infant Resuscitator fails any of these tests, the valve assembly should be regarded as faulty and replaced with a new valve assembly (see the parts list (section 4.1.2) for part numbers).
Constantine Yap
Membre depuis le 05/04/20
364 Réputation
2 tutoriels rédigés
Équipe
Cal Poly, Team S1-G6, Zimmerman Spring 2020 Membre de l'équipe Cal Poly, Team S1-G6, Zimmerman Spring 2020
CPSU-ZIMMERMAN-S20S1G6
4 membres
27 tutoriels rédigés
0 commentaires
Ajouter un commentaire
Nombre de vues :
Dernières 24 heures : 1
7 derniers jours : 3
30 derniers jours : 10
Total : 523