    • If the ventilator is not already in diagnostic mode, simultaneously press the ALARM RESET and 100% O2 keys on the front panel for approximately five seconds while turning ventilator power on, then touch OK to enter diagnostic mode. (Service Manual 8-19)
    • Reconnect the oxygen source to the ventilator. (Service Manual 8-19)
  1. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: étape 3, image 1 de 1
    • Touch Oxygen on the Hardware screen and set the flow to 1 LPM. Verify that the O2Position display reads from 185 to 525 steps. (Service Manual 8-19)
  2. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: étape 4, image 1 de 1
    • Touch Oxygen and set the flow to 180 LPM. Check that the O2 Position display corresponds to the set altitude: (Service Manual 8-19)
    • Connect a patient circuit and analyzer to the ventilator. (Service Manual 8-19)
    • Set the analyzer’s function to read LPM mode and gas select to O2 (set a high flow range capable of measuring 180 LPM oxygen). (Service Manual 8-19)
  3. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: étape 7, image 1 de 1
    • Check that set oxygen flow corresponds to measured flow: (Service Manual 8-19/20)
Ethan Norman
Membre depuis le 05/04/20
157 Réputation
1 tutoriel rédigé
Équipe
Cal Poly, Team S1-G4, Zimmerman Spring 2020 Membre de l'équipe Cal Poly, Team S1-G4, Zimmerman Spring 2020
CPSU-ZIMMERMAN-S20S1G4
5 membres
5 tutoriels rédigés
0 commentaires de tutoriel
Ajouter un commentaire
Nombre de vues :
Dernières 24 heures : 0
7 derniers jours : 0
30 derniers jours : 1
Total : 266